I would like to know if a specific seismometer can measure 1 micron/sec velocity. I have a few specs from the datasheet but I'm not a seismologist and am trying to figure out how to relate the specs to one another.

I have:

Velocity output band: 30s (0.03Hz) to 100 Hz
Output Sensitivity: 2400 V/m/s
Peak/Full scale output: Differential: +- 20V 
Sensor dynamic range: 137 dB @ 5 Hz

Thanks in advance!

$\endgroup$
1

2 Answers 2

1
$\begingroup$

This is a partial answer because I'm not an expert and because I don't know what the dynamic range of 137 dB means. Hopefully you can add a little more information.

tl;dr: if 137 dB is the dynamic range in power, then it's 68.5 dB in voltage and velocity which sounds more plausible, and makes the velocity sensitivity well below 1 micron per second. However we don't yet know what the noise and bandwidth of your signal are yet so we can't evaluate that.

I have a few specs from the datasheet...

The more information you share from the data sheet the better although I've now just noticed that the question is about three years old.

Also, there may be some helpful insight at How sensitive are typical seismometers?


137 dB is a factor of $10^{-137/10} = 10^{-13.7} =$ 2E-14.

Let's assume just for a moment that the dynamic range applies to voltage and velocity, rather than power.

If the full scale is 20 Volts then the minimum resolution (ignoring noise) is 4E-13 Volts.

If the sensitivity is 2400 V per m/s then it's

$$\frac{4 \times 10^{-13} \text{ V}}{2400 \text{V per m/s}} \approx 1.7 \times 10^{-16} \text{m/s} $$

which is $1 \times 10^{-10}$ microns per second, and absurdly low number.

On the other hand if the 137 dB is power then we multiply the 20 V by $10^{-137/\mathbf{20}} = 10^{-6.85} =$ 1.4E-7 to get a minimum resolution (again ignoring noise) is 2.8E-6 Volts, and

$$\frac{2.8 \times 10^{-6} \text{ V}}{2400 \text{V per m/s}} \approx 1.2 \times 10^{-9} \text{m/s} $$

which is $1.2 \times 10^{-3}$ microns per second or about 1 nanometer per second, which is indeed a realistic number for some capacitive and optical displacement sensors. It sounds plausible.

$\endgroup$
0
$\begingroup$

create a bench test using an acceleramentor and a transducer since their values are known then equate the results to your sesimo

$\endgroup$
1
  • 5
    $\begingroup$ Hi Chuck, welcome to stackexchange. This is a rather terse answer and, while it does offer a way forward, I think the asker of the question was hoping it was possible to tell from the datasheet. Would you consider expanding it and, if the answer cannot be determined from the datasheet, explaining why not? $\endgroup$ May 11, 2018 at 6:44